Fiat Coupe Forum
- Founded by Kayjey & James Northam
- Funded by the Club for the benefit of all owners
Fiat Coupe Club UK
join the club
Fiat Coupe Forum
 
» Announced
    Posting images


» Related sites
    Main club site
    fiatcoupe.net


» External data
    owners listed
 
Who's Online Now
2 registered members (Edinburgh, 1 invisible), 236 guests, and 3 spiders.
Key: Admin, Global Mod, Mod
Forum Statistics
Forums69
Topics113,596
Posts1,341,051
Members1,801
Most Online731
Jan 14th, 2020
Top Posters(All Time)
barnacle 33,553
stan 32,122
Theresa 23,300
PeteP 21,510
bockers 21,071
JimO 17,917
Nigel 17,367
Edinburgh 16,782
RSS Feeds
Club Events
Club Information
Track Events
Rolling Road/RWYB
Social Events
Non-UK Events
Coupé Related Chat
Coupé Spotting
Coupé News/Press
Buying/Selling Advice
Insuring a Coupé
Basic FAQ's
How to Guides
Forum Issues
Technical Problems
General Maintenance
Styling
Tuning
Handling
ICE and Alarm
Coupés for Sale
Coupés Wanted
Parts for Sale
Parts Wanted
Group Buys
Business Forum
Other Vehicles for Sale/Wanted
Other Items for Sale/Wanted
Haggling/Offers
Ebay links
Other Cars
Other Websites
General Chat
Previous Thread
Next Thread
Print Thread
Science lesson needed #1386968
23/10/2012 11:21
23/10/2012 11:21

E
Enforcer
Unregistered
Enforcer
Unregistered
E



Regarding my diabetic meter's control solution, I am confused.

The solution is specified as being 0.12% (by weight) glucose, which I think translates into 6.67 mmol/l. However, the margin of error allowed for the machine testing the control solution is given as 6.7 - 9.0 mmol/l , which is all above the value I just calculated. It's supposed to be an equal plus/minus tolerance.

Where am I going wrong here?

[PS The actual reading for the control solution is 8.1 mmol/l]


Last edited by Enforcer; 23/10/2012 11:35.
Re: Science lesson needed [Re: ] #1387076
23/10/2012 21:19
23/10/2012 21:19

E
Enforcer
Unregistered
Enforcer
Unregistered
E



Just to update:

I now have the control solution. It is described as a 0.12% glucose solution, which does equate to 6.67 mmol/l. However, I am told to be happy with a test reading of between 6.7 and 9.0, and indeed, it reads 8.1.

So what is going on here? The machine systematically reads high by around 20%. The manufacturers know this. But they don't tell me to adjust my blood glucose readings accordingly.

So all this time I have been reading my blood glucose as 6.6, when in fact it was about 5.3!

Re: Science lesson needed [Re: ] #1387124
23/10/2012 22:52
23/10/2012 22:52

E
Enforcer
Unregistered
Enforcer
Unregistered
E



When you walk,
through the storm,
hold your head up high,
and don't be afraid of the dark,
At the end of the ....

Re: Science lesson needed [Re: ] #1387131
23/10/2012 23:01
23/10/2012 23:01
Joined: Aug 2011
Posts: 2,595
angus, scotland
jimbob13 Offline
I need some sleep
jimbob13  Offline
I need some sleep

Joined: Aug 2011
Posts: 2,595
angus, scotland
....storm is a golden sky,
And the sweet silver song of the lark...


[Linked Image]
Re: Science lesson needed [Re: ] #1387133
23/10/2012 23:02
23/10/2012 23:02

E
Enforcer
Unregistered
Enforcer
Unregistered
E



Lonely in here, isn't it.

Re: Science lesson needed [Re: ] #1387134
23/10/2012 23:05
23/10/2012 23:05
Joined: Aug 2011
Posts: 2,595
angus, scotland
jimbob13 Offline
I need some sleep
jimbob13  Offline
I need some sleep

Joined: Aug 2011
Posts: 2,595
angus, scotland
I always wondered what purgatory would be like...


[Linked Image]
Re: Science lesson needed [Re: ] #1387211
24/10/2012 07:38
24/10/2012 07:38
Joined: Dec 2005
Posts: 33,553
Berlin
barnacle Offline
Club Member 18 - ex-Minister without Portfolio
barnacle  Offline
Club Member 18 - ex-Minister without Portfolio
Forum Demigod

Joined: Dec 2005
Posts: 33,553
Berlin
Try being a moderator...


[Linked Image]
Don't get no respect! Coupe Fiat 1994-2000 - an owner's guide <-- clicky!
Re: Science lesson needed [Re: ] #1387216
24/10/2012 07:50
24/10/2012 07:50
Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
M
MeanRedSpider Offline
Je suis un Coupé
MeanRedSpider  Offline
Je suis un Coupé
M

Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
Sorry - late to this. It's control solution not blood. Meters have to correct for all sorts of blood effects - haematocrit and other endogenous interferents. These don't exist within control solution. All you need to worry about is whether the meter reads within the specified range. It looks as though it reads nearly smack in the middle so that's good. Currently, under ISO standards, meter accuracy is only required to be (broadly) +/-20%. This is going to change.

Blood is incredibly complex stuff and varies wildly between individuals. This makes measuring glucose with a £15 meter pretty challenging.

Re: Science lesson needed [Re: ] #1387231
24/10/2012 10:13
24/10/2012 10:13

E
Enforcer
Unregistered
Enforcer
Unregistered
E



Right - so the reading for blood should be correct when the reading for a control solution is as stated - around 8.00 for a 6.7 solution. I'll make that assumption.

But then I still have a problem. When I went to the hospital to donate a blood specimen for measurement, I took my own readings immediately before and after the hospital visit. All within the space of forty minutes or so. The two readings I took measured 6.3 and 6.7, but the hospital reading was 5.3.

You might say that my FBG level is capable of rising and falling quite significantly, so no problem, but at that time I was never getting anything like 5.3 at home. It was a complete one-off. It is very difficult to believe that I would have measured 5.3 at home when the hospital reading was taken. Much easier to believe that there is a systematic bias of around 1 mmol/l on my machine.

So, the bottom line to all of this is simply that I would like to know how accurate my meter is. That's all. I want to be able to compensate for any systematic bias it might be introducing.

Let me reframe my question then:

Is it possible for a machine to give the predicted reading for a control solution, but still introduce systematic bias when testing my blood? Could there be something within the complexity of blood that throws it out?


Last edited by Enforcer; 24/10/2012 10:26.
Re: Science lesson needed [Re: ] #1387243
24/10/2012 10:57
24/10/2012 10:57
Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
M
MeanRedSpider Offline
Je suis un Coupé
MeanRedSpider  Offline
Je suis un Coupé
M

Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
In short - yes, absolutely it is possible. Haematocrit, for instance, is definitely an endogenous interferent that could affect the reading of a blood glucose meter. Ascorbic acid (Vit C), uric acid and a wide range of others can affect the results too - all of which are naturally occuring in your blood.

Forty minutes is a long time but you wouldn't expect such a change necessarily between those two. It's also unusual for the bias ("built-in" offset) of a meter system to be high.

We always advise against comparing meter readings between meter systems but only compare against a lab reference system (Yellow Springs Instrument or similar). Do you know what the hospital used to measure your blood glucose? It's very possible that they used just another meter in which case you can expect anything up to 30% difference (even 40% could be acceptable).

What you can say from the control solution test is that the meter/strip is functioning as designed.

Feel free to PM me if you prefer - happy to help either way

Re: Science lesson needed [Re: ] #1387246
24/10/2012 11:04
24/10/2012 11:04

E
Enforcer
Unregistered
Enforcer
Unregistered
E



Thanks -

The time between the hospital test and each of my home tests was only about 20 minutes, and also, as I say, at that time I was almost never getting anything under 6.00 at home.

The hospital is a major one (Glan Clwyd) and they had to measure all sorts of factors in my blood (lipids, HbA1c, total cholesterol, creatinine, etc., so I don't imagine they would have used a meter.

Re: Science lesson needed [Re: ] #1387258
24/10/2012 11:24
24/10/2012 11:24
Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
M
MeanRedSpider Offline
Je suis un Coupé
MeanRedSpider  Offline
Je suis un Coupé
M

Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
Originally Posted By: Enforcer
The hospital is a major one (Glan Clwyd) and they had to measure all sorts of factors in my blood (lipids, HbA1c, total cholesterol, creatinine, etc., so I don't imagine they would have used a meter.


Ironically they probably use one of "our" (different part of the company)blood testers. Interestingly, I don't know how accurate these are for glucose measurements relative to a proper reference system - I'll ask though. Were most of the other blood measurements well within "normal" ranges? (of course, you don't need to answer that). If they were, it's less likely that this is the cause of the effect you are seeing though doesn't discount it.

Agree - I don't think the time will explain the differences though your liver can release glucose into your blood stream in response to effort. Generally, though, you need to be slightly cautious of results separated by any more than a few minutes.

Are you still using the Boots system? If so, let me know the model and I'll see if I can find out if I have anything on its bias.

Re: Science lesson needed [Re: ] #1387295
24/10/2012 15:05
24/10/2012 15:05

E
Enforcer
Unregistered
Enforcer
Unregistered
E



http://www.journalofdst.org/September2012/PDF/VOL-6-5-ORG2-FRECKMANN.pdf

Accu-Chek Compact plus GT

Looks like any bias is minimal, and more towards the negative side.

I don't know how to explain the hospital experience. Either the hospital is using unreliable equipment or my machine is positively biased. Can you find out what they use at Glan Clwyd?

Re. My other results - cholesterol and LDL on the high side.

Last edited by Enforcer; 24/10/2012 15:08.
Re: Science lesson needed [Re: ] #1387304
24/10/2012 15:49
24/10/2012 15:49
Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
M
MeanRedSpider Offline
Je suis un Coupé
MeanRedSpider  Offline
Je suis un Coupé
M

Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
Need to be slightly careful with that particular study (Roche paid for it) but, yes, I expect your system is slightly negatively biased.

I'm finding out what to expect from hospital analysers.

By all means call Roche and ask them for an answer

Re: Science lesson needed [Re: ] #1387323
24/10/2012 17:51
24/10/2012 17:51
Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
M
MeanRedSpider Offline
Je suis un Coupé
MeanRedSpider  Offline
Je suis un Coupé
M

Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
So I have an answer and, to cut a very complicated story short, there is possibly a degree of inaccuracy in the hospital test too. It's very possible both systems meet their accuracy specifications but they just don't agree on the number.

Re: Science lesson needed [Re: MeanRedSpider] #1387324
24/10/2012 17:57
24/10/2012 17:57

E
Enforcer
Unregistered
Enforcer
Unregistered
E



Any idea which system they use, or the degree of uncertainty?

Re: Science lesson needed [Re: ] #1387400
25/10/2012 07:51
25/10/2012 07:51
Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
M
MeanRedSpider Offline
Je suis un Coupé
MeanRedSpider  Offline
Je suis un Coupé
M

Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
Originally Posted By: Enforcer
Any idea which system they use, or the degree of uncertainty?


Don't know what system they use but seemingly these systems could be as much as 10% out on glucose so it's possible that the difference on reading that you are seeing is shared between the two measurement systems - hospital lab and your meter

Re: Science lesson needed [Re: ] #1387482
25/10/2012 13:51
25/10/2012 13:51

E
Enforcer
Unregistered
Enforcer
Unregistered
E



Thanks a lot for all your help.

I'm discussing this with someone else on a diabetes forum, and after much thought and testing he tells me that his Freestyle machine reads 2.8 mmol/l control solution AS 2.8.

His other two machines read control solution AND blood too high, by the same margin. In all three cases, therefore, they do not discriminate between control solutions and real blood.

So as far as I can work out, the anomaly rests firmly with my machine. The control solution IS 6.67mmol/l, but the manufacturers specify a reading range of between 6.7 and 9.0 and, indeed, it does read as 8.1.

So either my machine discriminates in a unique way (unlike any other machine) between control solution and blood serum, or it just reads too high all round.

Last edited by Enforcer; 25/10/2012 13:52.
Re: Science lesson needed [Re: ] #1387573
25/10/2012 20:22
25/10/2012 20:22
Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
M
MeanRedSpider Offline
Je suis un Coupé
MeanRedSpider  Offline
Je suis un Coupé
M

Joined: Mar 2006
Posts: 8,852
Cambridge & Cotswolds
Our latest system automatically identifies control solution and deals with it slightly differently (it's pretty sophisticated in how it measures the blood)

I've replied to your PM and hopefully I'll have more answers tomorrow.

Do make sure you follow good practices: washing and thoroughly drying hands before testing, storing your strips in their container with the lid shut etc Given the relatively tiny quantities of blood used in these test, even the smallest quantity of contaminants causes errors

Re: Science lesson needed [Re: ] #1387583
25/10/2012 21:05
25/10/2012 21:05

E
Enforcer
Unregistered
Enforcer
Unregistered
E



OK - thanks!

BTW my machine works automatically with presealed drums of 17 test strips. They don't get contaminated.


Powered by UBB.threads™ PHP Forum Software 7.7.1
(Release build 20190129)
PHP: 7.3.33 Page Time: 0.013s Queries: 14 (0.006s) Memory: 0.8254 MB (Peak: 0.9736 MB) Data Comp: Off Server Time: 2024-04-25 19:32:26 UTC